Miscellaneous › Others › Analog signal chain – Two point calibration
Hi all
My question regarding the subjet is related to the expected error you can get after you done offset and Gain calibration.
I have a simple signal chain which include a PGA and 16 bit DeltaSigma ADC.
I can get very close values after calibration on each end of a ADC input range: 0 mv to 512 mv.
The corrected values after using the Corrected_Gain and offset are not very close to ideal Gain=1.
Near Full Scale the valus are good but the error increase down to cero volts.
Ther are error like INL which is +/- 2LSB , but the error I get is big compared to 2 LSB.
I would appreciate any help on that.
Regards,
you must take average of several codes….more will the number of samples less will be the gain error